decomposition products or even co-prescribed drugs include the most
common interferences. Thus, it is necessary to establish that the
chromatographic signal produced is only due to the analyte and not as a
result of the co-elution of other compounds. A bioanalytical method is
considered to be selective if the lack of response in the blank biological
matrix at the retention time of the analytes is demonstrated at the lower
limit of quantification (LLOQ). The determination of method sensitivity,
linearity, precision and accuracy is required by the three guidelines and
usually performed at three concentration levels in several replicates
(typically three to five). Linearity should be demonstrated within the
defined calibration range with the quality of the calibration curve as an
aspect of the greatest importance in bioanalytical method validation. In
fact, as Almeida et al. (2002) stated, the quality of bioanalytical data is
highly dependent on the quality of the calibration curve used to
extrapolate the analyte concentrations in unknown samples. The
calibration curve is the relationship between instrumental response and
the concentrations of the analyte and it must be generated for each
analyte, using a sufficient number of calibration standards (Table 2.1.5).
Precision, which expresses the closeness degree among individual
measures of an analyte when the same procedure is applied repeatedly
to multiple aliquots of a homogenous matrix, is frequently performed on
the same day (intra-day precision or repeatability) and over different
days (intermediate precision or inter-day precision), both expressed by
the relative standard deviation (RSD), which is the absolute value of
coefficient of variation (CV). Accuracy, which describes the closeness
between the mean experimental data obtained by the method and the
corresponding nominal concentration, is also determined intra- and
inter-daily and is expressed by the relative error (RE), as a percentage,
which is the ratio between experimental and nominal concentrations, or
by bias, which is the deviation from the nominal value as a percentage.
Matrix effects, carry-over and dilution integrity have been recently
integrated in the guideline issued by EMA as well as stability evaluation.
From a practical point of view, these parameters can be estimated from
the analysis of QC samples, which represent the future real samples.
Table 2.1.5: The ICH, FDA and EMA validation parameters and respective limits instituted.
Biomedical Chemistry: Current Trends and Developments
- Titel
- Biomedical Chemistry: Current Trends and Developments
- Autor
- Nuno Vale
- Verlag
- De Gruyter Open Ltd
- Datum
- 2016
- Sprache
- englisch
- Lizenz
- CC BY-NC-ND 4.0
- ISBN
- 978-3-11-046887-8
- Abmessungen
- 21.0 x 29.7 cm
- Seiten
- 427
- Schlagwörter
- Physical Sciences, Engineering and Technology, Chemistry, Organic Chemistry, Green Chemistry
- Kategorien
- Naturwissenschaften Chemie