3-DShapeRecoveryof theLeftHeartChamber from
BiplaneX-RayProjectionsUsingAnatomicalA-Priori
InformationLearnedfromCT
RolandSwoboda1,2, JosefScharinger2, andClemensSteinwender3
1 ResearchCenterHagenberg
University ofApplied SciencesUpper Austria,Austria
roland.swoboda@fh-hagenberg.at
2 DepartmentofComputationalPerception
JohannesKepler UniversityLinz,Austria
josef.scharinger@jku.at
3 Clinicof InternalMedicine I
GeneralHospital Linz,Austria
clemens.steinwender@akh.linz.at
Abstract
Recovering the 3-D shape of the left heart chamber from bi-planar 2-D x-ray projection images is a
challenging task sinceonly sparseandnoisydata isavailable for reconstruction. In thiswork,a3-D
statistical shape model (SSM) of the left ventricular (LV) anatomy is learned from high-resolution
CT data and utilized as a-priori information to solve the under-determined and ambiguous recon-
struction problem. A 2-D/3-D registration method fits the SSM to the x-ray images of the patient by
calculating simulated projections of the SSM and minimizing the difference between simulated and
given projections. The presented approach is evaluated using simulated and real patient data. For
patients where both projection images and CT data are available, the reconstructed LV is compared
to the true shape known from CT. Our results show a good correspondence between recovered and
trueshapes. UsingaSSMasanatomicala-priori information for reconstructionhelps in limiting the
spaceofpossible solutions andallows togenerate statisticallyplausible shapes.
1. Introduction
Cardiac diseases are one of the most common causes of death in the industrialized world today. In
the case of acute myocardial infarction, for instance, interventional x-ray angiography is state-of-
the-art for both treatment and diagnosis. To evaluate the viability of myocardium after infarction, a
catheter is advanced into the left heart chamber (ventricle) and contrast agent is injected to opacify
theventricular cavityduring radiation. Bi-planarcine-angiographicequipment isused toacquire two
x-ray image sequences simultaneously from standard right anterior oblique (RAO) and left anterior
oblique (LAO) views, seeFig. 1.