Sim.Metric #1 #2 #3 Mean Std.
dmin (mm) 2.43 2.32 2.95 2.57 0.34
dortho (mm) 2.36 2.05 3.36 2.59 0.68
DOV (%) 98.01 92.87 82.11 91.00 8.11
VOD (%) 74.72 80.13 68.12 74.32 6.01
κ (%) 87.49 90.41 79.75 85.88 5.51
Table2. EvaluationofLVshaperecovery fromthree in-vivoangiograms.
5. DiscussionandConclusion
In thiswork,anewmethodforrecoveringtheLVfromcontrast-enhancedbi-planarcine-angiographic
x-ray images has been proposed. The novelty of our approach is that a-priori information about the
LV anatomy is learned from high-resolution CT images, modeled as a SSM and utilized for recon-
struction. A2-D/3-D registration technique is applied to fit theSSMtoangiographicprojections.
When only two (noisy) projections are available, the reconstruction problem usually becomes under-
determinedandambiguous. Insuchcases, theincorporationofa-priori informationplaysanimportant
role, since this can limit the space of possible solutions and improve the ability to deal with noisy
data. In contrast to [7], anatomical a-priori information is derived from data of in-vivo instead of
post-mortem subjects; other approaches often do not utilize this kind of information at all. Although
onlyonebi-planaracquisition isused for reconstruction,ourapproach isgenerallynot limitedby the
number of projections. However, since additional acquisitions increase the amount of radiation and
bolus, thisnumber is usuallykept toaminimum.
Using a SSM for reconstruction allows to generate statistically plausible and patient specific shapes.
Unlike other 3-D LV SSMs often found in literature, anatomical areas like the apex, the atrial con-
cavity and the aortic valve region are preserved in our model. This is necessary to generate complete
contour and densitometric information; otherwise, additional errors are introduced in the reconstruc-
tion process. Further note that these areas typically overlap with the ventricular cavity in projection
imagesandare thereforehard to recover withoutprior knowledge.
Evaluation with both simulated data and real patient data shows promising results. The LV volume
is recovered at high accuracy. This is important for assessing volumetric diagnosis parameters, like
EF.Concerningshapeconformity, theoverlapbetweenoriginalandrecoveredvolumeishigh, though
there is still place for minor improvements. Future work will focus on improving the model fitting
processandon evaluating ourapproachwith more in-vivoangiograms.
References
[1] S. Benameur, M. Mignotte, S. Parent, H. Labelle, W. Skalli, and J. de Guise. 3D/2D Registra-
tion and Segmentation of Scoliotic Vertebrae using Statistical Models. Computerized Medical
ImagingandGraphics, 27:321–337,2003.
[2] T. F. Cootes, C. J. Taylor, D. H. Cooper, and J. Graham. Active Shape Models - Their Training
andApplication. ComputerVision and ImageUnderstanding, 61:38–59,1995.
51
Proceedings
OAGM & ARW Joint Workshop 2016 on "Computer Vision and Robotics“
- Title
- Proceedings
- Subtitle
- OAGM & ARW Joint Workshop 2016 on "Computer Vision and Robotics“
- Authors
- Peter M. Roth
- Kurt Niel
- Publisher
- Verlag der Technischen Universität Graz
- Location
- Wels
- Date
- 2017
- Language
- English
- License
- CC BY 4.0
- ISBN
- 978-3-85125-527-0
- Size
- 21.0 x 29.7 cm
- Pages
- 248
- Keywords
- Tagungsband
- Categories
- International
- Tagungsbände